Dear participant,
Thank you for your question. The test and validation image format, spacing, and resolution will fully resemble the training images. The 20x/40x notation can be confusing as it is scanner-specific, see the info below, thus we opt to use microns per pixel notation instead. The training, validation, and testing images are provided at a maximum resolution of ~0.25 microns per pixel (what is most widely referred to as 40x). The subsequent resolutions provided in the pyramidal tiff differ by a factor of 4. Thus each training, validation, and testing image has resolutions: 0.25, 1.0, 4.0, 16.0....
The intermediate resolutions (0.5, 2.0, 8.0) were removed from the images, to optimize the data storage/transfer/download.
Please reach out in case of any further questions.
Kind regards,
Khrystyna
Leica Aperio:
Scans at 20X, 40X or 63X (with oil immersion lens), have resolutions of 0.5, 0.25, 0.17 micron/pixel respectively.
https://med.virginia.edu/biomolecular-analysis-facility/services/shared-instrumentation/aperio-scanscope-slide-scanner/
Zeiss:
20x/0.8NA Plan-Apochromat (a=0.55 mm)
40x/0.95NA Plan-Apochromat (a=0.25 mm)
https://www.helmholtz-munich.de/pathology-and-tissue-analytics/zeiss-axio-scanz1/index.html
Phillips:
40x 0.25 mpp
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC8344580/
3DHistech P1000
40x 0.12 mpp
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC8344580/
KFBIO
20x 0.25 mpp
https://en.kfbio.cn/product/newProduct/subProduct/pro400/
Olympus
20x - 0.33mpp
https://www.uib.no/en/rg/mic/120747/olympus-vs120-s6-slide-scanner
Akoya
40x = 0.25 μm/pixel
https://www.akoyabio.com/phenoimager/instruments/