Expected Max Resolution at 40x in Validation/Test WSIs

Expected Max Resolution at 40x in Validation/Test WSIs  

  By: raja.fida on July 2, 2024, 5:36 a.m.

Hi all,

We're extracting 8k tiles at 40x from WSIs (resizing to 4k at 20x due to the unavailability of 20x). What’s the highest expected resolution at 40x for validation/test WSIs? This knowledge will help us determine the maximum number of tiles that could come from a WSI, and hence to ensure the proper running of our algorithm in the given computational constraints at validation and test time.

Thanks!

Re: Expected Max Resolution at 40x in Validation/Test WSIs  

  By: KhrystynaFaryna on July 3, 2024, 4:15 p.m.

Dear participant,

Thank you for your question. The test and validation image format, spacing, and resolution will fully resemble the training images. The 20x/40x notation can be confusing as it is scanner-specific, see the info below, thus we opt to use microns per pixel notation instead. The training, validation, and testing images are provided at a maximum resolution of ~0.25 microns per pixel (what is most widely referred to as 40x). The subsequent resolutions provided in the pyramidal tiff differ by a factor of 4. Thus each training, validation, and testing image has resolutions: 0.25, 1.0, 4.0, 16.0.... The intermediate resolutions (0.5, 2.0, 8.0) were removed from the images, to optimize the data storage/transfer/download. Please reach out in case of any further questions. Kind regards,

Khrystyna


Leica Aperio:

Scans at 20X, 40X or 63X (with oil immersion lens), have resolutions of 0.5, 0.25, 0.17 micron/pixel respectively.

https://med.virginia.edu/biomolecular-analysis-facility/services/shared-instrumentation/aperio-scanscope-slide-scanner/

Zeiss: 20x/0.8NA Plan-Apochromat (a=0.55 mm) 40x/0.95NA Plan-Apochromat (a=0.25 mm) https://www.helmholtz-munich.de/pathology-and-tissue-analytics/zeiss-axio-scanz1/index.html

Phillips: 40x 0.25 mpp https://www.ncbi.nlm.nih.gov/pmc/articles/PMC8344580/

3DHistech P1000 40x 0.12 mpp https://www.ncbi.nlm.nih.gov/pmc/articles/PMC8344580/

KFBIO

20x 0.25 mpp https://en.kfbio.cn/product/newProduct/subProduct/pro400/

Olympus

20x - 0.33mpp https://www.uib.no/en/rg/mic/120747/olympus-vs120-s6-slide-scanner

Akoya 40x = 0.25 μm/pixel https://www.akoyabio.com/phenoimager/instruments/

 Last edited by: KhrystynaFaryna on July 3, 2024, 4:16 p.m., edited 1 time in total.

Re: Expected Max Resolution at 40x in Validation/Test WSIs  

  By: kwatson on July 4, 2024, 11:28 a.m.

The resizing does not change the magnification. In fact, how do you know that 0.25 um/pixel is 40x? You don't... On many Aperio scanners, 20x is 0.25 um.px. It's also 0.5 um/px; there is no 40x, except in marketing brochures.